first of all I am a beginner with electrospray. Now I am using LC/MS with electrospray ionization in negative ion mode. I try to quantify some acidic carbohydrates with 0-3 sulfate groups/molecule.
First of all I don't want sulfates to drop off so I have tried to find mild conditions and still get some ionization. Addition to sulfate loss the molecules also tends to fragment readily.
I have used Agilent MSD low resolution detector. Eluent is 50mM Ammonia plus 20mM Ammoniumbicarbonate.
The main problem is that ionization seems to differ from day to day. The peak with most intensity can be 1)double charged, 2)singly charged with one sulfate loss. All the parameters are the same. I think that when I have higher concentration of analyte, the 2) comes predominant. I also get molecule with one charge and three charge. It would be ideal to ge singly charged only.
I have to smooth my raw data, because on low concentrations, chromatograms look awful, because ionization is so low. My lowest concentration is now 450pmol/ml but of course the column makes it more dilute. Flow rate is 0,5 ml/min and peak is 1 minute broad.
Please, can You tell me what to change to
1) get better ionization without loss of sulfates or further fragmentation
2) to get same ionization pattern every day, preferentially singly charged
3) to get same ionization intensity every day
4) what else I should think
Thank you,
sahram