HPLC calibration in normal phase

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi All,

I am working on a normal phase HPLC method from last 1 month. Now my HPLC is due for calibration.I have to now prepare my system for reverse phase for calibration purpose & then again convert it back to normal phase. Can anybody help me know is there any suitable method of calibrating HPLC system using normal phase mobile phase? Is it correct to use a system continuously in normal phase which has been calibrated using reverse phase methods?

thanks,
Vaishali
My opinion, and I'm not a cGMP expert, is there's no need to requalify for normal phase, or when you convert back. Just run system suitability each time you use it, and document that.

If your company has an SOP that says you must requalify each time you change, and you do cGMP or GLP, then you must, or revise that SOP.

Many companies seem to get their underwear tied up in knots over their own over-stict paranoias.
2 posts Page 1 of 1

Who is online

In total there are 52 users online :: 0 registered, 0 hidden and 52 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 52 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry