Empower 2: Ring Buffer Overflow
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
4 posts Page 1 of 1
We have an agilent 1100 with an RI detector that keeps getting this message around the second injection of a sequence. Anybody scene this one before?
Thanks
I've seen it with A1100s before - it doesn't seem to interfere with anything, has caused me no problems. I'm not sure waht causes it.
Thanks,
DR
It causes our instruments to shutdown when we get it
You've checked the A1100s for local errors, table sizes of projects etc.?
-How much RAM in the LAC/E?
-Does anything show up in the event logs on the LAC/E at the time(s) that the buffer errors show up?
-Is A1100 hooked up via Equinox card or Ethernet?
-If Ethernet, what is typical sampling rate and wave range for DAD runs?
-How many instruments are on this LAC/E?
-Has Waters been able to help at all (have you searched their www site for the error)?
Thanks,
DR
4 posts Page 1 of 1