Empower 2: Ring Buffer Overflow

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
We have an agilent 1100 with an RI detector that keeps getting this message around the second injection of a sequence. Anybody scene this one before?

Thanks
I've seen it with A1100s before - it doesn't seem to interfere with anything, has caused me no problems. I'm not sure waht causes it.
Thanks,
DR
Image
It causes our instruments to shutdown when we get it
You've checked the A1100s for local errors, table sizes of projects etc.?

-How much RAM in the LAC/E?
-Does anything show up in the event logs on the LAC/E at the time(s) that the buffer errors show up?
-Is A1100 hooked up via Equinox card or Ethernet?
-If Ethernet, what is typical sampling rate and wave range for DAD runs?
-How many instruments are on this LAC/E?
-Has Waters been able to help at all (have you searched their www site for the error)?
Thanks,
DR
Image
4 posts Page 1 of 1

Who is online

In total there are 37 users online :: 0 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry