This is the explanation from the application chemist:
How to create a calibration:
I. First data processing
1. One of a higher standard has to be processed with a “normal” processing method (Baseline, Peak find, Library search and Calculation of Area and Hight)
2. In the processed Sample (Acquired sample) enter in the Peaktable the column “Group” (right mouse button -> Column -> Group.
3. Sort the Peaktable by Name (right mouse button -> sort)
4. In everyline where you find one of your compound enter a “g” in the column “Group”
5. Check the Name and the spectra. If it´s not the right one open the “Hit table”
6. Is the compound in the Hit table, select the whole line, right mouse button -> selct hit. Now it is entered in the Peak table.
II. Create a Calibration
1. Go to the Calibration and create a new one (right mouse button -> new) To rename again right mouse button and “rename”
2. Open the window a) Calibration Table b) Standard Table c) Calibration Curve d) Spectra and e)Chromatogram
3. In the Calibration Table add the processed standard (right mouse -> add standard)
4. Enter again the column “Group” and sort by Group.
5. Delete all the other compounds, which are not relevant for the Calibration. (Minus button on the top)
6. Enter the right quantification mass or masses
7. Enter the concentration in the last column (the column is named by the sample name)
8. To calculate enter again right mouse button -> calculate standard.
9. Add the other unprocessed concentration levels. Right mouse -> add standard.
10. Enter the conentrations in the cloumns. You can use the “fill option” to do this. Select the column -> right mouse -> fill.
11. Calculate the standards (right mouse -> calculate standard)
12. After the color changed from green, red to black, the calibration is calculated.
III. Quantify a sample
1. Create a new processing method
2. Only select “ apply calibration” and add the calibration in the processing method
3. Now select the sample and process it.