I received this email from my distributor some time ago.
Agilent Technologies does not recommend the use of a UPS or other isolation device to supply power to
a GC, GC-MS, or ICP-MS system. The Agilent system creates a highly varying load to these isolation
devices. The oven of a GC, for example, is controlled by either turning the power on or off each cycle of
the line. This creates a varying load condition on the UPS or isolation device of approximately 100 Watts
when the oven heater is off and 2000 Watts when the oven heater is on. Since the heater may be
powered off for one cycle and powered on for the next cycle of the line voltage, this creates a highly
repetitive level of load variation. Many isolation devices require several line cycles to recover from high
load variations. These isolation devices do not have sufficient damping and capacity to correct for the
level and frequency of load variation created by the Agilent devices.
Some isolation devices create undamped oscillations when subjected to repeated cycle-by-cycle, highlevel
load variations. These units will actually create very high levels of transient voltage supplied to the
load. These undamped oscillations often exceed the specifications for the Agilent system or the isolation
device. This may cause early or unexpected failure of the Agilent Technologies system, the isolation
device, or both.
Additionally, the high variation in load is created at a 50/60 Hz rate. Use of UPS or other isolation
devices utilizing Ferro-resonant transformers, regulators, or supplies is specifically NOT
RECOMMENDED since the large load variations at the resonant frequency of the transformer may cause
the transformer to operate unpredictably. This operation often causes the Ferro-resonant transformer to
overheat or malfunction.
Please contact the manufacturer of the UPS or other isolation device if you wish to use these to supply
power to Agilent Technologies GC, GC-MS, or ICP-MS system. Inform the isolation device manufacturer
of the varying cycle-by-cycle load behavior of the Agilent system to obtain a recommendation from the
manufacturer for the proper size, capacity, and type of UPS or other isolation device.
I know of a 7890 system that operates on a UPS not far from where I work that has blown 2 FID circuit boards in just under 2 years. That UPS could probably be an area of concern. I guess the take-home point from this is to inform the UPS manufacturer/distributor about your equipment when purchasing one.