My theory at the moment is once the sample has separated on-column into the the keto and enol components, some tautomerism of those components then occurs. To explore this I'm altering the residence time on the column via flow rate, but would also manipulate the kinetics of the tautomerism by running at both 15 oC and 40 oC (plus the two different flow rates).
The Waters column care guide gives the temp range of 20-50 oC for "low pH in order to enhance selectivity". As I'm not worried about peak shape or resolution (at this stage), my only concern is physically damaging the column by running at < 20 oC. I can find plenty of info regards column damage at elevated temp, but nothing about the opposite. I'll gradually drop the temp so won't thermally shock it, however is shrinkage of the packing material at low temps a concern??