What is the most contributing cause amongst following in the huge solvent peak?

1. Leaking injector
2. High split ratio
3. Very low sample concentration
4. low temperature at injector
5. Dirty liner
4.

If you mean too large of an injection for #3, then 3 is correct as well, otherwise, just #4.

If #5 was a VERY dirty injection port, lots of junk coating everything, then yes to #5 as well.

Most common cause is a blocked splitter or splitter valve not open at inject time.

best wishes,

Rod
It can also be a very low flow rate due to either blockage or operator error.
Don't forget liner "overloading".
You could aso try opening the split vent sooner.
I fixed it. I was using low solvent void time. I increased it and it got fixed. Thank you very much for all your efforts.
5 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry