Column- XBridge C18 5 μm x 150 mm x 4.6 mm (tried three different gel lots) Columns all additionally work fine on other systems.
Mobile Phase A 90% HPLC Grade Water/ 10% LC/MS grade ACN + 0.1% TFA
Mobile Phase B 100% ACN + 0.08% TFA
Gradient- 1 min hold 100% ACN. 1-36 minutes 80%A/20%B to 5%A/ 95% B-hold at B and then return to starting condition.
Water and TFA are JT Baker and ACN is Fisher.
Flow rate 0.3 ml/min
Runtime 60 min
Injection volume -25uL
Detection UV at 220 nm
I have tried cleaning on the instrument using water, ACN, 30% Phosphoric Acid ( no column on), 50%/50% ACN/Water. Nothing seems to work. I have taken off the sinkers cleaned them and even run it with no sinkers on just to see what I could see and the blob was still there.
I ran 12 zero volumes with the gradient going through lines A and B yesterday and 12 zero volumes with the gradient going through lines C and D just to see if the contaimination is in one of the lines. I saw the blob regardless of what lines it is going through.
I have HPLC specific glass containers I use to ensure that there is no contamination coming from the bottles. I have run the same mobile phase, in the same bottles, on a different instrument and see no blob.
PDA spectra is seeing peaks in the 320-360nm range.
I could really use some advice! I appreciate your help in advance.
Kelly