Sample can not return to baseline for SD test
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
Why can't some samples return to baseline when I do the simulated distribution test? However the peaks sepreated very well. What's wrong with it? Do I need to change some prameters, like column flow, temperature program? I used 10 meters column and when I run D2887+PW655, I got the perfect sepctrum under the same condition.
-
- Posts: 584
- Joined: Mon Sep 13, 2004 8:22 pm
What happens if you do a blank run? If you just get a LARGE base line change, then your column needs to be conditioned. If you get peaks at the high end, then you probably have a contaminated inlet. What exactley do you mean by "when I run D2887+PW655", as D2887 is a method for up to C40 and Polywax 655 is used for methods going up to C90.
Gasman
Gasman
2 posts
Page 1 of 1
Who is online
In total there are 87 users online :: 1 registered, 0 hidden and 86 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Semrush [Bot] and 86 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Semrush [Bot] and 86 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science