SPME fibre stuck in septum issue

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi, has anyone come across the issue of getting Supelco SPME fibres stuck in headspace vial septa? We are using a CTC-PAL setup on a Varian GC and found that fibres have been caught in the septa, presumably as the protective needle is withdrawing from the sample vial.

Does anyone have any suggestions as to how we could check for needle penetration given that this normally happens in the heater shaker so isn't visible?

Could it be that we've just purchased a faulty batch of 100micron PDMS fibre assemblies?

Any advice would be gratefully received, many thanks in advance,

Sally
Odd. To check needle penetration take one of the needle where the fibre is already broken off, run a sample and see whether there is a hole all the way through the septum. If not, increase needle penetration.

Peter
Peter Apps
Thanks Peter, that's a great idea. I'll give it a go.

Sally
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry