Are there known negative consequences of constantly running an ECD at a higher temp? 330 degree C vs. 300 degree C for instance? This is not for the occassional bake-out but for use as a default setting.

We're kicking around the idea of running at a higher temp for PCB analysis because we've noticed a loss of sensitivity/recovery when running samples extracted from oil matrices vs. soils and waters.
I never liked to exceed the column temperature max.

I know that after oil analysis I ALWAYS had to run a bake followed by a solvent injection at regular temperature before the closing standard otherwise I would get low recovery.
queenann,

Perhaps something to consider; rather than raising the temperature, you could insulate the upper body of the detector (depending on the brand.)

Best regards,

AICMM
3 posts Page 1 of 1

Who is online

In total there are 35 users online :: 0 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry