A sustained temperature of 70°C will strip a hydrophilic, silane-based coating off most silicas, and will accelerate the dissolution of the silica itself in the presence of any base. That probably accounts for the high background you observed with ELSD with the Tosoh column. I don't see why it's necessary to use high temperature for such an application; we got a nice separation of PC, PE and PS (phosphatidylcholine, phosphatidylethanolamine, and phosphatidylserine, resp.) at room temperature using a PolyWAX LP column and a gradient from 95-50% ACN with 5 mM ammonium formate, pH 6.5, and with ELSD detection.
The Tosoh Amide-80 column does work well for sugars in general, but reducing sugars will exhibit anomer separation unless the mobile phase is made somewhat basic by inclusion of something like 0.1% triethylamine. Using an amino column, polymeric or otherwise, renders that unnecessary because the stationary phase provides the locally high pH necessary to accelerate the mutarotation that causes the anomer peaks to collapse into a singlet peak.