(whole day) to equliberate. What's the problem with it?
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Everything is ready except for rear flow/pressure, because this GC can install three columns and we only install one colume in middle. I just set front and rear injector pressure control as 0.1 psi and the front actual total flow is 10mL/min and the rear actual total flow is 25mL/min. Does anybody know what's the total flow mean? How can I solve this problem?I'm working on Varian 3800 as well, when I reboot it, it always takes a very long time
(whole day) to equliberate. What's the problem with it?
Plug the column connections on the unused inlets, and set the pressure to 1 psi so that the controller has something to do. Set total flow to 5 ml/min or so - that gas will come out through the split vents of the unused detectors.Everything is ready except for rear flow/pressure, because this GC can install three columns and we only install one colume in middle. I just set front and rear injector pressure control as 0.1 psi and the front actual total flow is 10mL/min and the rear actual total flow is 25mL/min. Does anybody know what's the total flow mean? How can I solve this problem?I'm working on Varian 3800 as well, when I reboot it, it always takes a very long time
(whole day) to equliberate. What's the problem with it?
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.