Help with Degasser

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hey guys,

I am a PhD student and I have a small question. I am working with a shimadzu series 10 HPLC that has a UniFlows Degasys DG1310 degasser.

I wish to clean or do the required maintenance for it, as I have been getting some ghost peaks in my gradient run lately and the size of the ghost peaks has reduced after bypassing the degasser. This degasser has never been maintained and I do not have a manual for it (couldnt find one on the internet as well).

Can someone help me with some basic diagnostics and cleaning procedures to ensure that the degasser still has some life.

Thanks

Mukul Ashtikar
you can do 2 things
wash with high organics
100% ACN or THF

you can also use very mild detergents as well.
do not use a pure solution!!
use a diluted solution whose Ph is above 2 and not more then 9, this can also help clean the degasser channels.
What you are doing already is what we have had to do in our industrial lab as our degasser is also broken and is actually introducing air rather than removing it! We just ensure we degas our mobile phases by filtration and sonication before attaching the solvent lines and you can also use an end line filter too (if you dont already) to minimise air then simliar to you plumbing them in bypassing the degasser. If you're doing this and leave mobile phases for more than a few days sitting perhaps re-sonicate to minimise any issues. We havent noticed and detrimental effects on our baselines from degassing like this to using the actual (working!) degasser. The cost of parts/engineers to fix whatever has gone wrong isn't really worth it unless you're really short of time to spend degassing.

lynz x
Thanks unmgvar and lynzjm for the comments...

@Lynzjm, the degasser is working it probably had some kind of microbial buildup or got contaminated somehow...

Anyway purging the degasser with 1% acetic acid and then with ACN seems to have done the job.

Cheers!

M
aaah thats much nicer to fix lol! We use 50/50 IPA to flush our equipment lines as well! Glad you got it sorted :) Wish ours would come back to life too!

lynz x
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry