HPLC or UHPLC?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
I work with a company which is trying to upgrade their Lab. They don’t have any analytical instrument at the moment. I have not worked with any UHPLC system is the past, but it seems that they are becoming more popular. I am wondering if it would be better to invest on a UHPLC system rather than a HPLC system! Our main work would be on development of stability indicating assays for semisolid dosage.
-
- Posts: 195
- Joined: Fri Oct 22, 2004 6:02 pm
Depending on the nature of analysis and budget limitqation, a "HPLC" system may satisfy your needs in most cases. However, it makes more sense to consider buying LC systems that are compatible with both UHPLC and HPLC applications. For your purpose, such UHPLC/HPLC system is appropriate. Agilent, Thermo and Waters are offering such systems with some differentiations. You will need to contact the local sales person from all three companies for more details.
Xiaodong Liu
-
- Posts: 16
- Joined: Mon Jun 27, 2011 6:25 am
You must consider your application/s, as there are many which are not (yet?) compatible with UHPLC.
- There are very few columns for large molecules.
- Refractive index (RI) detectors are not available for UHPLC (CAD is the only current replacement).
-In a regulated environment (USP, EP), going from HPLC to UHPLC is OK but not the the other way around.
- Do you have high throughput to justify UHPLC (maybe try semiporous/nonporus columns on HPLC).
In addition to what XL said most new "off the shelf" HPLCs can allow >600bar.
I am working with a HPLC/UHPLC hybrid and would not reccommend it as a starting kit.
In short: go for a good new HPLC and contact column supplier for applications for your analytes. Add UHPLC when throughput trumps costs.
- There are very few columns for large molecules.
- Refractive index (RI) detectors are not available for UHPLC (CAD is the only current replacement).
-In a regulated environment (USP, EP), going from HPLC to UHPLC is OK but not the the other way around.
- Do you have high throughput to justify UHPLC (maybe try semiporous/nonporus columns on HPLC).
In addition to what XL said most new "off the shelf" HPLCs can allow >600bar.
I am working with a HPLC/UHPLC hybrid and would not reccommend it as a starting kit.
In short: go for a good new HPLC and contact column supplier for applications for your analytes. Add UHPLC when throughput trumps costs.
3 posts
Page 1 of 1
Who is online
In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: satya5 and 29 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: satya5 and 29 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science