Carry over problem in Chloramphenicol analysis

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
i am working in a waters alliance 2695 HPLC coupled with quattro micro LCMSMS and masslynx software.In Chloramphenicol analysis we have facing a carry over problem.The HPLC coulmn we are used is agilent zorabax eclipse XDB-C18(3.1x150mm)and mobile phase is .1% formic acid in water in A and .1% formic acid in Acetonitrile in B and the gradiant pattern is 95% A is for 2min and from 2-3min. it is changed to 5% A it is hold for 3min. from6-7 it is changed to 95% A and maintained at the end ie. 12min. I am chaned run time to 15min. then also the carry over is observed at the same RT(5.11min.).The other thing i observed is the intenal standard is not carry overed.Anyone facing this type of carry over problem please help me


David Antony
Have you tried using a needle wash?
Are you using deuterated CAP as the internal standard? If so, are you sure the CAP peak is carryover, and not lab contamination? It makes no sense for CAP to carryover, but not dCAP (assuming that your concentrations are comparable, of course).
Prolonging the equilibration time won't help solve either a carryover or contamination problem. You can go back to using 12 min runs if your retention times were stable.
I assume you are running this in negative ion mode, in which case you also don't need formic acid in the mobile phase.
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
Dear David,
Most LCs have CO problem to some extend. You should observe CO for IS as well if it is in the vial and not automated addition from separate IS reservoir in the autosample (some configurations have this). Try washes and all suggested tricks. Reduce concentration; use fully caped Si or polymeric based columns. Run between every injection a quick run or 2 of a solvent to purge everything.
Alexandre
"If your experiment needs statistics, you ought to have done a better experiment." Rutherford
3 posts Page 1 of 1

Who is online

In total there are 53 users online :: 0 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 53 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry