problems between integrator and uv/vis detector

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Whenever I turn on the detector of the prep hplc, the integrator goes crazy and shows a whole lot of random peeks instead of a baseline. As soon as I turn off the detector, it shows a straight baseline. I have tried so many things but can't find the problem. I'm stuck and don't know what to do now. Any suggestions would be most appreciated.
If hooked up correctly to detector, make sure integrator attenuation isn't set way sensitive, maybe you're seeing big giant baseline noise. With detector off, there is no signal, even tiny.
I think you are right. I checked the integrator settings and it seems they are back to default. The attenuation was 0 and when I raised it I immediately saw a decrease in the baseline noise. Thanks so much.
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry