Hi,
For that type of splitter, not only do you get different RT, you also get different split ratios in a GC running a oven gradient programme. (less split into FID at higher temperatures)
as the ~.12 cm of .1mm column into the 280°C MS transfer line, is
not long enough to behave as a constant flow restrictor at typical flow rates, and the measured pressure (even in constant flow mode) at the split is different at low to high oven temperature programmes,
I can comment that Agilent sells their MicroFludic Splitter (MFS) system, that needs the EPC module so as to supply a source of Helium at constant pressure,
http://www.chem.agilent.com/Library/Sup ... 0E6FBE.pdf
Also, the EPC ensures there is enough Helium for the varying volumes of makeup Helium depending on the oven temperature. eg if memory serves me correctly, @ 40°C the MFS makeup was ~2.8 mL/Min, @ 280°C it was ~0.8 mL/Min (the makeup dilution adversely effected the MDL on both detectors, so I canned the MFS, but the split ratio was constant).
You could also try a twin hole ferrule for the injection port, and run dual columns, the RTs won't be the same, but as you will be using the FID for quantitation, you will only be reporting on peak areas for a particular RT.
Have you trying running your MS in SIM mode, as this may improve your precision (%RSD) as it creates more points under a peak, which is what I assume you are calling accuracy when stating (+/- 0.5% results).
Alex