Galaxie Chromatography Data system
I would appriciate if somebody can help me with the constructing calibration curve for the CP-3800 Varian GC using the Galaxie Chromatography data system.
I have oppened the new sequence and entered the 5 levels of the standard I am using, click for the first one :Clear old points and than add levels for the other up to 5th level.
Also on the method in the lefet bottom screen under the calibration, enetered the date and the name of the new calibration curve, and new 5 levels with the calculated ammount of my standard. Everything is saved, Gc run all samples but when wanted to open the calibration curve I can not.
What can I do, or what I am missing, I knew we can do something through reprocessing but do not know the right way.
Thank you
What kind of error message (if any) do you have when opening your calibration curve ?
You should contact Bruker for asking support as Varian GCs are now supported by Bruker.
Best regards
On the method page -under calibration.
Is the correct calibration curve file selected ?
I'm guess yes but worth a check.
If your working on the chromatogram are you saving
Chromatogram -the local copy
or chromato method - the library copy.
A screen shot or two of what is happening would be useful.
Thank you, I have solved the problem I had with the calibration curve.
Savka