Hi,

I was wondering why cant we analyse water samples in GCMS?

Will water spoil the column or the detector(MS).

Thanks,
Jason
Depending on the operating temperatures, water can oxidize both the column and parts of the MS. Oxidation in the MS leads to poor tuning and dramatic loss in sensitivity. If you are running a cooler oven gradient (i.e., <100°C or so) water may not damage a good methyl-silane too much. Even using a good purge-and-trap system, some water makes it through the GC and into the MS. I'm not sure how much per injection it would be on a µL scale but it's actually quite a bit. The MS (Agilent 5973) seems to do fairly well requiring a source cleaning only about every 6 months while running samples about 18 hrs a day. Columns would last about 6 months too.

If you are going to run volatile compounds (e.g., B.P. < that of napthalene) then I would just suggest using a purge and trap concentrator system for water analysis. It is far more sensitive than you will ever get by injecting a microliter of a water sample.

Hope this helps a little.

Ty
~Ty~
2 posts Page 1 of 1

Who is online

In total there are 41 users online :: 0 registered, 0 hidden and 41 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 41 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry