- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
OK, supervisor says we should just dilute samples and standards further so thatany such "impurities" are hidden in the HPLC baseline and do not meet 3:1 signal to noise, and re-validate the method. But that really does absolutely nothing to improve the signal to noise of the API v. the fragrance component. Does this make any sense to anyone, like is this an accepted technique? And I assume that I must revalidate the test procedure if I choose to inject 2.5 ul instead of 5 ul (I estimate if the peak was halved that we couldn't see it at all), that's what supervisor said.
So I responded to schedule in which month I should do this, as other work would be suspended during that. Our other fragrances in same matrix do not have the same tiny component in their fragrances; we sure would hate to have a separate validated test procedure for just one fragrance, would cause confusion at QC locations. Comments appreciated, thanks.