I am trying to determine diethyl phthalate (and other phthalates eventually) by GCMS.
I have run a blank ramp with no injection and got no 149 peaks.
I have inserted the clean syringe needle and got some 149 peaks.
When I manually inject clean hexane (determined by fid on another system) I get 149 peaks too.
The GC-MS system gets more septa particles inside the liner than the fid system.
The GC-MS is Saturn 2000 with 1079 injector.
How do you go about setting up a GC-MS for this tricky analysis?
Do you bake out the phthalates from the septa? Do you use a duckbill?
Any help will be greatfully appreciated.
(I've tried HPLC-UV - too many co-elutions!)
WK

Phthalates contamination have many sources. My suggestion is to keep all plastic and rubber products away from the instrument and samples. This includes laboratory gloves, pipette bulbs, natural rubber vial septa, ect....

Make sure to use a PTFE lined vial septa. You may also want to avoid septa on the autosampler "wash" vials.

Wrap all septa in aluminum foil if necessary.
Answer Man
2 posts Page 1 of 1

Who is online

In total there are 460 users online :: 0 registered, 0 hidden and 460 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 460 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry