I'm working with an Agilent 7890 GC coupled to an Agilent 7000 Triple Quad MS. I'm validating a method for pesticides (organochlorines and organophosphates). I run one replicate at a time, from the lowest to the highest level, and then the second replicate the same way, and so on. All injections are from the same sample but different vial. My problem is that I don't get good precision injecting that way. I don't know what causes this. I put 600 uL solution in each vial (2 mL capacity), I control the temperature, I check the autotune parameters and everything is ok. I noticed when I plot the results for each replicate I get excelent r2 coefficients with linear fit (>0.999), but when I plot all the replies together the r2 coefficient drops to values such as 0.95. Furthermore, I realised that in general the highest level of the curve has higher variability (this is the latest level injected for each curve replicate). The run time is ~35 min.
What could be causing this? Should I change my injection sequence?