Detection of Ethanol using Purge and Trap
Posted: Thu Aug 04, 2011 8:44 pm
by jeffsmitheverybody
We have a client that is looking to have us run analysis for Ethanol at a detection limit of 400 ug/L. We were looking to see if anyone has tried Purge and Trap and if they have made any changes from the typical 820B method?
Are you still using a #10 OI trap (tenax/silica gel)
Any help would be greatly appreciated.
Re: Detection of Ethanol using Purge and Trap
Posted: Fri Aug 05, 2011 12:59 am
by Karen01
Never mind ... I got my units mixed up
Posted: Fri Aug 05, 2011 7:00 pm
Update:
I am currently working on a #7 Trap. Was able to achieve the 400ug/L in Scan mode. Now working on getting lower detection limits.
I am going to run with the following operation conditions
sample temp 60C
11 min purge at 40C
2 min desorb at 190C
standard water management settings
I am also going to try to run in SIM mode as the retention time has been established using Scan mode.
Posted: Mon Aug 08, 2011 5:14 pm
by LindseyPyron
You should be able to hit that level easily. See an application note below:
www.estanalytical.com/presentations/California Oxygenates and 8260 App Note.pdf