Detection of Ethanol using Purge and Trap

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello,

We have a client that is looking to have us run analysis for Ethanol at a detection limit of 400 ug/L. We were looking to see if anyone has tried Purge and Trap and if they have made any changes from the typical 820B method?

Are you still using a #10 OI trap (tenax/silica gel)

Any help would be greatly appreciated.
Never mind ... I got my units mixed up
Update:

I am currently working on a #7 Trap. Was able to achieve the 400ug/L in Scan mode. Now working on getting lower detection limits.

I am going to run with the following operation conditions

sample temp 60C
11 min purge at 40C
2 min desorb at 190C
standard water management settings

I am also going to try to run in SIM mode as the retention time has been established using Scan mode.
You should be able to hit that level easily. See an application note below:

www.estanalytical.com/presentations/California Oxygenates and 8260 App Note.pdf
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry