Recently I got a specific peak shape problem during RP chromatography and using MS/MS (QqQ) detection. Tried couple of weeks to troubleshoot it without success. Any of your insights and comments are welcomed.
Firstly, about the hardware: Agilent 1100 coupled with API 4000 QTrap
Column: 3.0 mm ID, C18 column with an appropriate guard column
Mobile phase: methanol/THF/10mM formate buffer pH3 (isocratic elution)
Folw: 0.4mL/min, t(column) = 30 degrees
Ionization mode: ESI
The picture of the SRM chromatogram with the problematic peaks (designated with arrows and numbers):

All compounds are nitro-containing compounds (proprietary) with molecular weights below 250.
They all have similar backbone (same groups - arranged differently on the same backbone -> positional isomers) and the problematic peaks No. 2, 3 and 4 have additional OH (hydroxyl group), while the peak No. 1 has additional COOH (carboxyl group). They all show severe fronting and tailing, resulting in a very broadened low intensity peaks. Other peaks are symmetrical and do not show the same problem (as you can see from the picture).
So far: - I changed the column and its guard column with new ones (no reslut)
- I changed the tubing. From pump to the injector i have SS tubing, while from injector to column and from column to ESI source - PEEK tubing. Reduced the length of the pre and post column tubings (no result) as much as it is allowed.
- made PM, changed the injector rotor seal and pump seals. Made the leak and pressure tests afterward. Everything seems to work properly with all tests passed
- i did passivization of HPLC according to the Agilent procedure with nitric acid (because the peak No. 4 cannot be detected, which usually gives nice peak)
- increased the buffer ionic strength twice - the problem still persists.
Under the same conditions only on Thermo HPLC and MS the chromatography is nice, without fronting/tailing and sensitivity issues.
I ran out of ideas what to change next in order to improve the overall picture. I suspect that the problem is before the column, not after it, but maybe i am wrong. The MS was recently PM-ed (calibrated , cleaned and works normally)
Any ideas?