JA » Thu Mar 31, 2005 9:07 pm
chsp,
The buffers you've listed are all carbonyl-containing and are known to absorb at low-UV wavelengths where you are working. As Provetech mentioned, the absorbance of your MP at the end of the gradient is less than the initial conditions due, mostly, to the decrease in concentration of your buffer additive.
Using an additive in line A and an organic solvent in line B is a rough and ready means of buffering a system. If you have a multi channel low pressure mixing system you can minimise the baseline deflection when using salts by utilising 3 channels, for eg. A=H2O, B=ACN and C=aqueous buffer. Modify your gradient to draw a constant 10% of line C and form a gradient using A and B. Prepare your buffer at 10x the desired concentration (eg. 0.1 M). This idea is reproduced from many Waters application notes I've seen.
If you're just using the acids, you can add them directly to both the Water and Acetonitrile.