I have already connected pegasus III to computer via com1 serial cable for comunication and optical cable for data collection and looks fine. I don't see any error on the logfile.
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
I think you meant "The GC does have a serial cable from the GC to the computer. The computer's COM1 goes to the Pegasus and COM2 goes to the GC"The GC does have a serial cable from the GG to the computer. The computer's COM1 goes to the Pegasus and COM2 goes to the computer. I would strongly suggest you contact LECO service and talk with them. My experience has been that I can call and talk to a live person - and they will privide you with a description of what you need , part numbers, and prices. And, they can answer other questions for you as well.
You were correct on my mistyping where the com port is located.
With the tune issue, the messages indicate that the instrument was not able to locate either a sufficiently strong signal for PFTBA or there was no signal at all. Did you have a profile displayed during optimization? If so did it show a peak for m/z 69. Also, what was the vacuum reading?
Yes I had profile display but it was a flat red line at zero count (1636800 zero of vertical axis and 68 zero of horizantal axis). There was no peak at all. All I see a flat red line. Vacuum reading is 5e-7 torr
A very important question: Was the instrument operating at the time you purchased it? An instrument that has been improperly stored or shipped has increased potential for problems.
No. Instrument was brand new but sitting in the warehouse for years and never been used or turned on as the owner said.
A few questions:
1- default setting of Ion source temp. is 180 deg C. I followed exactly according to the manual and it does not say anything about this tempreature. Is that ok to leave it on 180 Deg C or change it? In case of change what is the best temp.?
2- I haven't installed transfer line yet. Status bar shows 400 deg C for it. Could this cause any problem for now in terms of optimization?
3- There are 10 fuse boxes on the GP unreg/ssr board (to tyhe right) with a one LED on each of them . 6 in the right boxes (black fuse boxes: labeled U1-U6) and 4 in the left boxes (red fuse boxes: labeled U7-U10). When instrument is ready all black fuse boxes LEDs are on except U5 which starts blinking when I connect transfer line heater momentarily. All red fuse boxes LEDs are off except U8 which is always on. I think this also should be off if evrything is ok. Problem should be something related to this LED. Am I right?
About email to LECO I emailed them once and they asked for the instrument serial number at the back label but there is no SN number at the back label. I sent them a shot of the back label. That is probably why I haven't heard from them yet.
.