I‘ doing HS- GC method development of residual solvents, ethylmethylketon(MEK) and 1- propanol.
The RSDs for peak area of the two analytes from 5 injections of the standard solution are not bad (about 2%). But the customer will have to raise the sample volume (1000 µl instead 100 µl). Unfortunately, the RSD by 1000µl was terrible (8%). Should I raise the agitator temperature or extend the incubation time?
Method parameters:
Diluent: Water
Standard Solution: MEK :0,03 mg/ml
1-Propanol: 0,1 mg/ml
FID Detector.
Inlet temp.: 250°C; Detector temp.:300°C
Oven temp.: 40°C for 6 min., 15°C/min to 300°C
Column: DB 1 (30*0,32mm*1µm)
Carrier gas: Helium, Flow rate: 1,9 ml/min, Split ratio: 10 to 1
Injection volume: 1 ml
HS: Agitator: 80°C, Incubation time: 30 min. Syringe temp. 90°C
HS-Vial size: 10 ml and I’m putting 100 µl into the HS vial.
Thanks in advance.
Tin sang