why so many peaks in gradient run

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
hi guys,

When ever i am using gradient program i always used to find lot of peaks.
not only at the sarting and ending of graient program.
EXAMPLE: I have a mixture of highly polar, medium polar and non polar compounds to separate

A:Buffer (KH2PO4)
B:ACN or methanol
Time %A %B
0 100 0

10 100 0

30 20 80

50 20 80

60 100 0

65 STOP

problem is these small gradient peaks interfering my non polar impurities at 0.1% Level
and some times not reproducible.

Do u think anyof these following reasons

1.instrument make
2.Gradient ramp
3.Solvent/Buffer grades
4.Premixing of organic to buffer.

PLEASE ADVISE
Regards
Krish
This is one of the more common questions that gets asked on the Forum. Watch this video for a discussion of what's going on and a diagnostic test to confirm the cause of the problem (you will have to register separately).
http://www.lcresources.com/more_resourc ... hp?f=3&t=5
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry