Recovery in Headspace GC
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
I am facing problem of recovery of the residual solvents spiked with the sample at low level for a drug whereas I find the recovery of about 80-85% when spiked at specification level of the solvents. Can somebody suggest me how to approach this problem?
-
- Posts: 48
- Joined: Fri Sep 24, 2004 8:21 pm
hi, what about your matrix? GC conditions? and tell me what kind of volatiles compounds are spiked? are you on the EPA oe USP?
Thanks!
Thanks!
-
- Posts: 8
- Joined: Sat Sep 25, 2004 10:17 am
Hi,
The matrix is a bulk drug buffered with sodium carbonate. Of the various solvents the problem is only with the recovey of methanol. All other solvents are perfectly getting recovered.
The matrix is a bulk drug buffered with sodium carbonate. Of the various solvents the problem is only with the recovey of methanol. All other solvents are perfectly getting recovered.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
It is difficult to trouble shoot with so little information.
methanol is fairly volatile but hydrophilic compared to many other solvents. Adding salts to an aqueous sample does not always improve recovery as much as may be desired.
Reducing sample size can improve recovery of methanol and other hydrophilic analytes.
Methanol also can be adsorbed to active sites in your system. (flatter response)
One more problem is that it can be difficult to plug focus if your injection volume is large. (flatter response)
Sorry I cannot be of more assistance.
methanol is fairly volatile but hydrophilic compared to many other solvents. Adding salts to an aqueous sample does not always improve recovery as much as may be desired.
Reducing sample size can improve recovery of methanol and other hydrophilic analytes.
Methanol also can be adsorbed to active sites in your system. (flatter response)
One more problem is that it can be difficult to plug focus if your injection volume is large. (flatter response)
Sorry I cannot be of more assistance.
-
- Posts: 8
- Joined: Sat Sep 25, 2004 10:17 am
Thank you chromatographer1. Your suggestions helped us a lot. The problem got solved by reducing the sample size.
5 posts
Page 1 of 1
Who is online
In total there are 33 users online :: 1 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 32 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science