Shimadzu 2010 retention time moving..
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
while doing shimadzu 2010 auto injector calibration iam getting different retention times and areas for 200 micro litre injection.we are having 500 micro litre loop..could you solve.....
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
first check the pump actually
if the flow is not stable it will give you not only retention time drifts but also area drifts
as the peak retains more the area increases, slower flow higher area
if the flow is not stable it will give you not only retention time drifts but also area drifts
as the peak retains more the area increases, slower flow higher area
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Check the dead time (t0).
If dead time and retention time (tR) change by the same percentage, then you have a flow problem.
If they both change by the same absolute amount, then there is an issue with the the start signal timing (perhaps the autosampler cycle time?).
If the dead time stays constant and retention time changes, then you have a chemistry or temperature problem.
If dead time and retention time (tR) change by the same percentage, then you have a flow problem.
If they both change by the same absolute amount, then there is an issue with the the start signal timing (perhaps the autosampler cycle time?).
If the dead time stays constant and retention time changes, then you have a chemistry or temperature problem.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 1
- Joined: Sat Jul 16, 2011 2:03 am
Hi!
Please check if the rotor seal has been replaced when you have installed the 500 ul loop.There is different HPV rotor seal(with bigger grooves).
If this has not been replaced and the old one(with small groove) being used, you have no problems upto 100 ul injection.But when you increase the volumemore, then the area and RT variation occure!
Cheers!
Rahul
Please check if the rotor seal has been replaced when you have installed the 500 ul loop.There is different HPV rotor seal(with bigger grooves).
If this has not been replaced and the old one(with small groove) being used, you have no problems upto 100 ul injection.But when you increase the volumemore, then the area and RT variation occure!
Cheers!
Rahul
4 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Hollow, Semrush [Bot] and 19 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Hollow, Semrush [Bot] and 19 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science