HPLC DAD FL noisy baseline

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

21 posts Page 1 of 2
.
What was the mobile phase you were working with?

Are IPA, acetone 7 methanol compaible with all of the MP components.

What happens if you try an aqueous wash?
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
.
.
Flow without column gives the same noise as with column and with no flow with or without column? All solvents give the same noise?
Did you try to push solvents, even air, through the cell with a syringe?
What HPLC system are you using? Agilent systems include built-in tests for lamp intensity, flow cell transmittance, etc.

Poor sensitivity and noisey baseline sound like very little energy is getting to the detector. As you have suggested this is either a dirty flow cell or weak lamp. I would change one or the other. Instrument manufacturers have cleaning procedures for UV cells. Some use nitric acid but I dont recall the details.

What wavelength are you monitoring? Do the problems change at a different wavelength? Collect all spectra during a run and if there any structure in the data you either have a dirty flow cell or your solvents are contaminated.

Good luck!
A. Carl Sanchez
.
How did the baseline look during "manual" push of solvents through the cell? Just wondering whether you got some column stat. phase fines in the cell.
By the way, what do you mean with "the solvent bottle did run a LITTLE dry"?
And: did you do a stop flow or not? You didn´t do the baseline tests without a column on line?
.
Well one doesn´t do a no flow without having flow first, then compating to stopped flow. There are some questions you didn´t answer.
.
does the noise change when you remove the cell?
A. Carl Sanchez
.
It would be a strange coincidence if the lamp went bad just when you did non-LC with an LC. I don´t know your particular instrument and just found out from your last statement that FL stands for fluorescence. If this is correct and both use the same cell then you have another flow path (especially cell) testing possibility. Also different solvents and flow rates are helpful, and, furthermore, the use of a syringe for pushing solvents through.
My guess remains: Column fines in the cell.
21 posts Page 1 of 2

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry