tlahren » Wed Jul 13, 2011 1:43 pm
I don't think the MDH would be able to help over seas or even out of state. As far as proficiency testing you might try
http://www.accustandard.com. You can order a PT sample and analyze it. They post the results online and I think they also mail them to you with your score (how many analytes you correctly quantitated). They also carry pesticide standards for standard methods.
For GC/MS try starting with the U.S. EPA method 8270. It's at least a good reference for GC/MS analysis of semi-volatile analytes. The list for this method contains numerous pesticides and the quality check criteria included a check for degradation of DDT (must be very low) and peak tailing for pentachlorophenol and benzidine (also needs to be low). AccuStandard also has most EPA methods available for download at
http://www.accustandard.com/asi/epa_downloads.php3. I'm not familiar with the CIPAC methods but the EPA 8270 method has enough info to get you down the right path of putting out good, reliable data. I'm not sure which is the method for LC/MS but I think it's an 8300 number. There are many so take a look at the link above.
You are right in that it can take months for method development for each method on each instrument and that proficiency must be shown on each specific instrument (2 different GC/MS systems must both be tested for same method proficiency). The info in the EPA methods usually has a basic GC/MS setup scheme complete with GC and MS parameters that should get you at least seeing and identifying compounds in your first couple analyses. Also ordering mixes of standards from vendors like AccuStandard, Restek etc., they usually will provide you with a chromatogram of the standard along with the analytical parameters and conditions in which they analyzed it. They will go one step further if you call them and probably chat with you about how to optimize methods for specific analytes as well. They are very helpful during method development usually.
Hope this further helps your goals.