Alp » Wed Jul 13, 2011 7:19 am
I would doubt your columns are bleeding off silica. The mobile phase you are using, 75/25 ACN/Aq 0.1% formic acid will prevent the silica from dissolving. Unless your water is basic (>pH

I would not be afraid of running without any formic acid as most columns work well in the range of pH 2 to 8.
Dissolution of silica becomes a problem above pH 8.
The agilent website gives some info on this column
http://www.chem.agilent.com/en-US/Produ ... fault.aspx
None of the listed methods use formic acid in the mobile phase, it is all ACN/Water.
http://www.chem.agilent.com/en-US/Produ ... 58220.aspx
This page also states the pH range the column can work with is 2 to 8 , so a little formic acid should not hurt at all.
Unless the column you are using has been subject to non-recommended mobile phases, I would doubt it will cause you a problem with white ppt.
I have seen mobile phases containing ACN give a white or biege PPT in the source. You can test if this is happening in your case by just running moblephase for a day or two, with no column attached.
A method for sugars by lcms is found at
http://mass-spec.stanford.edu/assets/20 ... omeaux.pdf
It also uses an amino column and ACN/Water.
Alp