Axar wrote:
I guess, best way to determine residual solvents is use of HeadSpase.
Yes, it is highly effective.
I would use water as solvent (in a case you don't need to determine solvents, insoluble in water, such as hexane, coz you will have problems with standart solution).
There is no problem in doing std addition. Put all solvents in a carrier solvent such as DMF or DMAc. I have done solvents such as hexane, methylene chloride, methanol, dioxane, ethanol at the same time, same sample.
If you water sample solution is 200 microliters, then add 50 microliters of DMF containing the required amounts of solvents for which you are testing. Add the same amount of DMF to each sample but change the solvent content for the std addition requirements
It is not necessary to dissolve your substance complitely. A dosage of UV will drain organics from your substance. Adding solts, such as Na2SO4 is option. In a case of water, it will boost evaporation for polar organics, and decrease it for nonpolar.
IT IS REQUIRED to dissolve the sample completely. Your analysis of solvent content will be inaccurate otherwise. Adding salts can be useful but almost always in NOT required, unless you use too large a sample size WHICH IS NOT RECOMMENDED. UV light? Why would one use such a variability technique which could produce decomposition products?
I do not think you need an internal standart in this case. I have reprodused a lot of residual solvents methods, and I have met internal standart rarely in them.
Rodney George