Changing columns between GC & GCMS?
Posted: Tue Jun 28, 2011 5:37 pm
by SmileyJ
I have GC MS data using a HP-5ms column and now I'm trying to get data on GC-EAG-FID. I'm wondering if I can install the HP-5ms column I was using on a HP GC-MS and put it in a Shimadzu GC-17a GC-FID?
Is it bad to change a column from a GC/MS to GC/FID?
Re: Changing columns between GC & GCMS?
Posted: Tue Jun 28, 2011 6:24 pm
by tlahren
No it's not bad to use the same column on two different detectors. Some applications use MS and FID simultaneously using a Y splitter from one main column. Usually the "ms" columns have lower bleed (of siloxane peaks) in an MS chromatogram. Shouldn't make a difference in an FID.
Re: Changing columns between GC & GCMS?
Posted: Tue Jun 28, 2011 7:26 pm
by SmileyJ
I thought so. Just never had a chance to use multiple different instruments.
Re: Changing columns between GC & GCMS?
Posted: Wed Jun 29, 2011 1:20 pm
by AICMM
SmileyJ,
Don't expect the retention times to be the same since the vacuum of the MS does the chromatography at the end of the column and this does not happen on the FID.
Curiosity killed the cat here.... What is GC-EAG-FID? Specifically, the EAG??
Best regards,
AICMM