Changing columns between GC & GCMS?

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
I have GC MS data using a HP-5ms column and now I'm trying to get data on GC-EAG-FID. I'm wondering if I can install the HP-5ms column I was using on a HP GC-MS and put it in a Shimadzu GC-17a GC-FID?

Is it bad to change a column from a GC/MS to GC/FID?
No it's not bad to use the same column on two different detectors. Some applications use MS and FID simultaneously using a Y splitter from one main column. Usually the "ms" columns have lower bleed (of siloxane peaks) in an MS chromatogram. Shouldn't make a difference in an FID.
~Ty~
I thought so. Just never had a chance to use multiple different instruments.
SmileyJ,

Don't expect the retention times to be the same since the vacuum of the MS does the chromatography at the end of the column and this does not happen on the FID.

Curiosity killed the cat here.... What is GC-EAG-FID? Specifically, the EAG??

Best regards,

AICMM
4 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry