Thanks for the quick reply. The MS system hasn't been tuned for a while, that is true, but I am not sure this is the problem here. Let me elaborate...
When I use the same source (and housing) and solutions on an ion trap MS that we also have in the lab, I get the expected fragment ions in both APCI+ and APCI- (tune parameters between QQQ and IT were more or less the same). I use only acetic acid as an additive in the mobile phase, however, I don't see any adduct ions in the full scan MS (not from the additive, solvent or from other possible contaminants such as sodium, etc.). Moreover, when I run Q1 full scan and Q3 full scan, the resulting spectra look more or less the same (abundant precursor ion). The collision gas readout equals the set value of 1.5 mtorr (Ar), so I am running out of ideas what I could be doing wrong (or what is wrong with the MS system).
Oh, and another thing, when I use ESI+,- MS2 or SRM with other compounds, I don't seem to have any problems with observing product ions, so could this by some strange coincidence also be ion source-related?
If there are any additional ideas I would be happy to try them out!
Thanks...