Pressure fluctuate proplem ??
Hi there ,
I am using this pumb
when i start my analysis the pubm read ( 26 mPa ) then it rise up to ( 30 mPa ) then reduce to ( 29 then 32 then 27 mPa ) .
I usually degas my mobile phase and my samples are just water with analyte .
Does this will effect my peak area ??how can i solve the problem
my column ( 250 mm * 4.6 mm ) , flow rate = 1.2 ml/min
- Purge the pump head to make sure that no air bubble is trapped.
- If this does not help a pump head maintenace might be helpful.
Your pump seems to have a piston backflush system. Make sure that the solution in there is exchanged frequently.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Hello. Air bubble is a probability problem. Also-small leaks. Look, there are no puddles.
But in general, jumps can be. Pump- one plunger. The plunger goes back and forth, and the balls old valve may not have time to practice.
Need a damper - a long thin tube of milliliter internal volumes .
What problem can be strong-noise baseline. For higher concentrations it is not critical