Dear all..
I have left my c18 column in 100% methanol after the end of my work today.
Tomorrow, I need to start my samples with 93:7 water:Acetonitrile.

So how should i change my eluting solvent ..?
Can I directly go to the 93/7 (Water/ACN) system or should i follow some procedure

with best wishes
HPLC System: HPG system with two(2) Knauer S1000 pumps- Dyamic Mixing Chamber-Knauer s3900 auto sampler-S2300 RI detector-S2500 UV detector- Raytest radioactivity detector

tarzan
Hi tarzan
I never do such big steps from 100 % organic to > 90 % aqueous. Instead I would first go from 100 % MeOH to 100 % ACN and then add water in 20 % steps, at least 2 column volumes each step. As you are using two pump system this is straightforward. But if you are using isocratic runs with 93/7 % v/v, pre-mixed buffer will give you much better baseline stability, especially when using an RI detector.

Kind regards
Jörg
Water and methanol are perfectly miscible, so I wouldn't expect any major problems. Do take in to account that certain water-methanol mixtures have a viscosity higher than either water or methanol, so watch out for over-pressure!
hi.
Thanks for the quick replies ..

Any ways .. how do you guys leave your columns at the end of the day ...

100% organic phase ..ie either methanol/ethanol/ACN

or

in certain composition of organic/water ...?

with best wishes
HPLC System: HPG system with two(2) Knauer S1000 pumps- Dyamic Mixing Chamber-Knauer s3900 auto sampler-S2300 RI detector-S2500 UV detector- Raytest radioactivity detector

tarzan
I always revert to 100% ACN with a small flow rate (0.005mL/min) on my C18 UPLC column.
It'll be interesting to see what others do.

Since methanol, ACN and water are all completely miscible in all proportions, and assuming this is a modern C18 column (i.e. more-or-less indestructable) you can change it between any of these samples as quickly as you want with no risk, except as LiVD said, there will be a pressure rise on passing between methanol and water because of 50% mix being more viscous than either alone.

I leave my C18 columns in highish organic, either 70-80% MeOH or ACN. The only reason for not going 100% is a personal, and not very logical panic, about gradual precipitation of anything salty. Although column-abuse is not to be encouraged, modern simple reverse phase columns are incredibly robust.
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry