Shimadzu LCMS 2020 Cone Voltage

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello,

I'm attempting to get a mass of some metal clusters by ESI on our Shimadzu LCMS-2020. The literature suggests low cone voltages. I've been unable to locate a cone voltage parameter, or determine the equivalent Shimadzu name. Does anyone have any experience with this?
The Shimadzu LCMS has a Q-array voltage you can change from the autotune value which is set as default in the MS tab of your method details page. Try setting a lower value for further details call there support at 800-477-1227
Hi,
The equivalant to cone voltage is the DL (desolvation line) bias. By default it is set to 0. If yoy infuse your compound or perform FIA you can automatically optimize this paramter to have the highest signal for your cluster.

Kind regards,
3 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry