Basically, this really only applies to everyone in a regulated industry.
What I am after here is what do we feel needs to be done to transfer a method from a HPLC to a UPLC, given the current draft revision of USP <621>, where linearity velocity takes into account column particle size. I'll defend that any day to an FDA auditor, no matter if the USP ends up accepting it or not. We must remember that the USP is a guideline to be followed, and that we do not have to follow the USP to the letter as long as we can scientifically justify our actions.
Having said that, what is necessary for demonstrating a transfer? I would say linearity, LOQ/LOD, and specificity due to different detector types and the general acceptance that not all C18 columns, for example, are created equal.
That amount of work would be only if you wanted to play it safe, which I would assume most QA departments would request.
But really, what are your thoughts? I'm really after plain and simple benchmarking here...