Hello,

Installed on HP5890, our 7673A controller and / or turret keeps stopping during an injection.

On starting a sequence, The turret initialises, the arm picks the vial from the tray, dispenses the vial into turret vial holder. The arm then withdraws and the turret tray rotates slightly , though the vial never actually reaches the sampling position.

After a few seconds, the gripper arm will move and then try again to pick up the now missing vial. If you do not now replace the original vial (currently in the turret) back to its original autosampler tray postition, you get - ERROR MESSAGE 'no vial in gripper'.

However, if you do replace the same vial from the turret back to the autosampler tray, the arm will again pick it up, transfer to the turret as before and after a few seconds you get ERROR MESSAGE 'Error 157' and controller repeatedly flashes red 4 times.

Any advice welcomed,

Lee
first thing... is the nut that hold the turret down loose? Sometimes that can loosen can cause the turret to fault
first thing... is the nut that hold the turret down loose? Sometimes that can loosen can cause the turret to fault
There is no nut on the turret for a 7673A, the turret just pushes down and snaps into place. But we would typically get the message "turret error" if that was out of place.

Also remember to take apart and wipe clean (no lubricants) the guide rods in the autosampler arm, maybe every two years. Can also wipe clean the rods in the tower.
Have you had any power cuts recently? Has the autosampler been turned on and off? You may need to reinstall the firmware for it.
cheers
Liz
first thing... is the nut that hold the turret down loose? Sometimes that can loosen can cause the turret to fault
There is no nut on the turret for a 7673A, the turret just pushes down and snaps into place. But we would typically get the message "turret error" if that was out of place.

Also remember to take apart and wipe clean (no lubricants) the guide rods in the autosampler arm, maybe every two years. Can also wipe clean the rods in the tower.

I'm using a HP 5890 w a 7673 autosampler. I am consistently getting a turret error. I have placed the turret on the injector turret motor but it doesn't seem to snap into place, this model does not have the nut on the turret. When the injector goes through the start up procedure, the turret seems to be quite loose and the rotation of the turret seems to be hampered by the guide arm in the tower. How much force is required to get the turret to 'snap' into place.
If the shaft and turret are lined up correctly, I've always found that it does take fore force than "expected" to seat the turret.
6 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry