krickos » Mon Jun 20, 2011 3:12 pm
Hi
Have not encountered your problem myself, but Ag can interfer with EDTA titrations. In titrations you typically control the pH to reduce interfering ions and you can add masking agents (nasty ones) such as CN- (cyanide) or F-.
See page 7 for instance in this link:
http://www.bhsu.edu/Portals/91/Analytic ... pter13.pdf
Technically speaking, if your pH electrode leach Ag ions or inner eletrolye from a sleeve or similar it may interfer during a longer pH measurement, at least in theory.
If you are opreating on very low pH values close to 1 you may have the "acidic error" that cases a drift upwards.
For "sleeve" saltbridge/junctions issues make sure that the electrode is properly maintained and calibrated link:
http://downloads.vertmarkets.com/files/ ... ph_wol.pdf
Usually you allow some electrolyte to flow out of the junction to prevent "attack on electrolyte or the Ag/AgCl wire", loww on electrolyte and you will have issues. Normally you have high purity KCl as electrolyte, usually can only interfere at very high pH, if its not so pure it may of course have interfering contaminats.