Ghost peak bigger after the GC has been in standby
Discussions about GC and other "gas phase" separation techniques.
7 posts
Page 1 of 1
We have a brand new GC with an FID and have a ghost peak around .8 minutes after the air peak. We are only injecting humidified air. We are trying to figure out why we are getting this ghost peak. The peak becomes bigger after the GC has been in standby for a while, compared to in a run that quickly follows the previous one. The peak almost disappears after 5 to 10 repetitive air injections that are made every 20 seconds or so. We also see this ghost peak if we use ONLY our calibration cylinder that is water free. Does anyone have any ideas?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
The forum archives are a great resource for ghost peak and carryover troubleshooting. It sounds to me as if your peak is a contaminant rather than a ghost.
Peter
Peter
Peter Apps
-
- Posts: 199
- Joined: Mon Sep 22, 2008 7:50 pm
Maybe a contaminant in the carrier gas that's building up on the head of the column between runs? More time in standby = bigger peak.
-
- Posts: 2
- Joined: Wed Jun 08, 2011 1:06 pm
The initial peak height and width is approximately the same after in standby (as little as 2 minutes) as compared to days of standby. With repetitive injections very close together (~20 seconds), the ghost peak disappears for the most point. Let it sit 2 minutes and we are back in full force. Let it sit a day, same initial ghost peak intensity as the 2 minute standby. I don't think it is the carrier gas since we are using two different carrier gases (automatic and manual injections). Has anyone had any problems with contaminants coming from the injection valves? If so, what could be done? Our problem is this ghost peak isn't going away even after tens of thousands of injections. You would think the intensity would decrease with usage of the injection valve. Very confused at this point.
-
- Posts: 50
- Joined: Wed Jun 08, 2011 11:34 am
It is not a carrier gas for sure. I guess it could be a contaminant in inlet. Try to clean liner (if you have one), septum and syringle.
All I ever need to know I'm learned in cloning vats.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
There is an active site in the injector/column that collects an impurity which volatizes in the carrier gas. When the active site becomes saturated, the contamination reaches maximum concentration as a plug. Otherwise it doesn't get focused but only becomes a continuous column bleed artifact.
A good GC is like an operating room. clean clean clean. And it isn't always easy.
best wishes,
Rod
A good GC is like an operating room. clean clean clean. And it isn't always easy.
best wishes,
Rod
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
You have a "brand new" GC and yet you have done "tens of thousands of injections"
To have different carrier gasses for manual and automatic injections sounds like a rather odd setup - cetainly not the run of the mill split splitles inlet with liquid injections. What hardware do you have, and how are you using, and what are you analysing ?
Peter
To have different carrier gasses for manual and automatic injections sounds like a rather odd setup - cetainly not the run of the mill split splitles inlet with liquid injections. What hardware do you have, and how are you using, and what are you analysing ?
Peter
Peter Apps
7 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science