GC-TOF-MS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
We have recently purchased a GC-TOF-MS (Waters GCT) I am developing a testmix to assess its performance. I will be using Peak Assemetry as one factor, on Quads we use S/N in the TIC. Using the same testmix it fails the S/N of one compound, however I feel this is irrelevant for this instrument. Would the Peak Area of an Extracted Ion be more appropriate?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
What the parameters have to be for a test mix, and the composition of the test mix itself, should be guided by the real analyses that you want the instrument to do. If you are doing quantitative trace analyses then LOD or LOQ (which you would probably measure as S:N for a small peak) is a useful parameter, if you are trying to "identify" trace componenets then you need to know what the smallest peak is that you can pull a full scan spectrum off. If your samples are full of acids, the shape of an amine peak might not be particualrly relevant.
Peter
Peter
Peter Apps
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
In evaluating the S/N ratio a few other thoughts: If I recall correctly, the Waters GCT is an accurate mass instrument. It acquires data a bit differently than a unit mass instrument - such as the typical quadropole. If you can select the accurate mass range to be included in an ion trace, the signal to noise ratio may vary, even on an ion trace, depending on the width whether you are including a width of 1 amu or some fraction thereof.
Also consider the sampling rate of the instrument and any smoothing that happens in the software during processing.
Also consider the sampling rate of the instrument and any smoothing that happens in the software during processing.
-
- Posts: 12
- Joined: Tue May 25, 2010 11:41 pm
Its for qualatative analysis of unknown samples containing phosphorous compounds.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
For such a specific application you need a special test mixture - preferably one that contains compounds as close as possible in structure and properties to the ones that you are looking for.
Think of all the things that could compromise performance for the application, e.g. poor S:N, poor chromatographic resolution, poor mass accuracy and resolution, thermal degradation of sensitive compounds, and check for those.
Peter
Think of all the things that could compromise performance for the application, e.g. poor S:N, poor chromatographic resolution, poor mass accuracy and resolution, thermal degradation of sensitive compounds, and check for those.
Peter
Peter Apps
5 posts
Page 1 of 1
Who is online
In total there are 65 users online :: 0 registered, 0 hidden and 65 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 65 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 65 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science