Column/solvent compatibility

Posted: Wed May 11, 2011 8:58 pm
by eric3
Hi,

I have a DB-5.625 column [(5%-Phenyl)-methylpolysiloxane]
in my 6890/5973 GC/MS. Is acetonitrile compatible with this column? What about DMSO or methanol? I usually run all standards and extracts in methylene chloride, but a possible client would like their sample run in acetonitrile, if possible. I would appreciate any advice for this newbie.

Thanks

eric

Re: Column/solvent compatibility

Posted: Wed May 11, 2011 10:20 pm
by chromatographer1
There should be no problem with the solvents listed.

However, be aware that DMSO especially can have OTHER chemicals dissolved in it. Likewise, methanol may have a large amount of water. The unknowns and the water can be damaging to the column phase and surface.

best wishes,

Rod

Re: Column/solvent compatibility

Posted: Thu May 12, 2011 11:17 am
by willnatalie
I've only noticed problems with acetic/basic materials. I would say mix a small amount with a solvent you normally use, inject it and see if you get a good peak from it. If the peak is good then you should be fine.

Re: Column/solvent compatibility

Posted: Thu May 12, 2011 12:13 pm
by eric3
Thanks everyone!