What eaxctly is the reason that carrier gas in GC must be pure?
Sandra
A couple of reasons come to mind right away:

If contaminants can be detected, they add to the detector signal For example, a small quantity of methane (a possible contaminant in helium, which is obtained as a byproduct of natural gas production) would increase the background on and FID.

If conaminants can condense in cool spots in the GC, they will form peaks. So, if you have hydrocarbons from a pump oil present in the carrier gas they will collect on the column when it is cool. When you do a temperature programmed run, the column heats up and the various compunds begin to partition into the gas phase as the temperature rises - resulting in the presence of peaks, even with no injection being made.
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry